Reading: Selection of Weighing Balance and Weight for Calibration
Inst Tools > Blog > Calibration > Selection of Weighing Balance and Weight for Calibration
In this article, you will learn the selection procedure of weighing balance and weight for calibration purpose.
Last updated: September 28, 2022 3:24 pm
Saurabh Vaidya
Calibration
2 Min Read
Selection of weight for Weights Calibration
A standard weight should be a higher class of accuracy than the weight to be calibrated.
Selection of weight for Weights CalibrationSelection of Weighing Balance for Weights CalibrationSelection of Weight for Weighing Balance Calibration
E1 class is the highest accuracy and M 3 class is the lowest accuracy.
Uncertainty of calibration of the Standard weights used will be 1/3 of its specified maximum permissible error.
Class of Weights that can be Calibrated
For calibration of E2 class weight, the E1 class should be used as standard.
For calibration of F1 class weight, the E2 class should be used as standard.
For Calibration of E1 class weight, standard weight should have similar or better metrological characteristics like magnetic properties, surface roughness, corrosion, etc.
Standard weights of each class shall have the technical requirements such as shape, construction, material, marking, etc as per OIML-R111-1.
 M1, M2, and M3 class weights are not recommended to be used as a reference for the calibration of weights. However, M1 weights can be used for above 20kg with coarser uncertainty.
In the above table, the Maximum Standard deviation of repeatability allowed to calibrate a specific class of weight is provided.
Selection of Weight
Let’s take an example
Weighing Balance of maximum capacity of 200 g with resolution d = 0.001 g
n is capacity divided by resolution
n= 220/0.001 = 220000
now check in the above table, where this n value fits.
n value fits in the 2nd column.
I.e., for the Weighing Balance of the maximum capacity of 200 g with a resolution of 0.001 g, it should be calibrated by E1 or E2 class weights.
Accuracy of standard weight used for calibration of weighing balance should be minimum of 1/3rd of readability (1/3 of least count of balance)
Reference Document OIML-R-111-1 and OIML R-76-1
DP Diaphragm Capillary Seal Level instrument Calibration Procedure
Differential Pressure Switch Calibration Procedure
Sample Raw DataSheet for Pressure Calibration
How to Test a Pressure Gauge
RTD Calibration
DeadWeight Tester Questions
Pig Signalling Switches Functional Testing
Stability and Uniformity of Temperature Bath
Capacitance Level Sensor Principle, Limitations, Installation & Calibration
Fieldbus Flow Transmitter Configuration
How to Calibrate pH Electrode?
Level Transmitter Calibration Procedure
How to Test a Pressure Switch ?
Instrument Calibration Record
How to Create Calibration Records?
Calibration Instruments
Basics of Calibration
Closed-tank-DP-Level-Transmitter-with-wet-leg-Zero-elevation-Calibration
Closed tank DP Level Transmitter with wet leg Zero elevation Calibration
Technologies for Oxygen Gas Measurement
How to Calibrate Oxygen Analyzer? – O2 Sensor Testing Procedure
Magnehelic Gauge Calibration
How to do calibration checks of vibration Probe, extension cable and vibration monitor
Smart Transmitter Calibration Tutorial
Smart Transmitter Calibration Tutorial Part 1
Working of Air Compressor
Working of Air Compressor – Types – Control Techniques
Van Stone Type Thermowell
Types of Thermowells
PID Controller Loop Tuning Questions and Answers
PID Controller Loop Tuning Questions and Answers – Part 2
thermocouple types
What is a Thermocouple ?
Synchronous Machine Quiz
Four Wire Current Loop
4-wire Transmitters Current Loops
Membrane Bioreactor in Water Treatment - MBR Process
Membrane Bioreactor in Water Treatment – MBR Process
Facts of Control Valves
Facts About Control Valves