Inst ToolsInst ToolsInst Tools
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Search
  • Books
  • Software
  • Projects
  • Process
  • Tools
  • Basics
  • Formula
  • Power Plant
  • Root Cause Analysis
  • Electrical Basics
  • Animation
  • Standards
  • 4-20 mA Course
  • Siemens PLC Course
Reading: Pressure Transmitter Preventive Maintenance
Share
Font ResizerAa
Inst ToolsInst Tools
Font ResizerAa
  • Courses
  • Design
  • PLC
  • Interview
  • Control System
Search
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Inst Tools > Blog > Pressure Measurement > Pressure Transmitter Preventive Maintenance

Pressure Transmitter Preventive Maintenance

In this article, you will learn the pressure transmitter and sensor calibration and preventive maintenance (PM) procedures.

Last updated: June 3, 2022 11:29 am
Editorial Staff
Pressure Measurement
1 Comment
Share
6 Min Read
SHARE

In this article, you will learn the pressure transmitter and sensor calibration and preventive maintenance (PM) procedures.

Contents
Pressure Transmitter Preventive Maintenance ProcedurePressure Transmitter Checklist

Perform the below-mentioned activities before starting the PM job:

Take proper work permit with the specific tag number of the pressure transmitter.

Following Test and Calibration Instruments (TCI) are required for the maintenance activity.

  • Digital Multimeter
  • Scandura or Pressure Pump as per the range of pressure transmitter
  • HART communicator
  • Tool bag with all necessary tools

Test and Calibration Instruments to be checked one day before the PM job.

Make sure the battery-powered instruments are properly charged.

Below mentioned calibration sticker (or similar kind of tag) is to be verified on TCI to confirm the validity of the instrument’s calibration.

calibration sticker

All mandatory personal protective equipment (PPE) is to be arranged before starting the maintenance job.

Check the pressure transmitters datasheet and note down the transmitter measuring range.

Pressure Transmitter Preventive Maintenance Procedure

Pressure Transmitter Preventive Maintenance
  1. Take appropriate work permit and take necessary approvals from the concerned persons.
  2. Bypass the logic if the pressure transmitter is having any interlocks (Always take bypass approval from operations and fill the Interlock Bypass Form).
  3. If a pressure transmitter (PT) is used for process control using a control valve, then ask the operation engineer to put the PID controller (control loop) in manual mode.
  4. Ask the operation engineer to arrange for the isolation of the pressure transmitter through a process isolation valve. (if needed)
  5. Verify the pressure transmitter tag number. You should always ask the value shown by the pressure transmitter on the DCS/SCADA/PLC systems and cross-check the value on the transmitter installed in the field. This is a double confirmation for identifying the proper tag. (optional)
  6. Note down the value shown by the pressure transmitter before starting the PM job.
  7. Open the back cover of the pressure transmitter and you will find the cable connections and terminals.
  8. Note down all the cable connections. So that after job completion, all cable connections are properly restored back.
  9. Check the supply voltage.
  10. Connect the HART communicator to the pressure transmitter.
  11. Confirm that the pressure transmitter is isolated with the process tapping point.
  12. Isolate and check the zero of the pressure transmitter. Open the equalization valve and
  13. Properly drain the fluid if needed and check the zero reading (for hazardous service close drain if required).
  14. Connect the Scandura or Pressure Pump or Pressure Source to the pressure transmitter.
  15. Cross-check the range from the datasheet and the pressure transmitter configured range.
  16. Apply the required pressure to the pressure transmitter with the help of a pressure pump or pressure source.
  17. Note down the pressure transmitter reading in the field and also note down the control room reading (DCS/PLC/SCADA).
  18. Repeat the above step for different pressure points like 0%, 25%, 50%, 75%, and 100% of the pressure transmitter configured range. Log the values in the below table.
% of RangeInput from TCIValue on field displayValue on DCS/SCADA/PLC
00  
252.5  
505  
757.5  
10010  
757.5  
505  
252.5  
00  

Note: Above table shows a pressure transmitter range of 0 to 10 kg/cm2

Keep checking for leakages with Snoop liquid leak detector (or any other solution) after increasing pressure at regular intervals.

Pressure Transmitter Calibration Setup
Pressure transmitter calibration and maintenance activity in the field
  1. If the pressure transmitter generates the output with an error less than the maximum permissible error, then both readings are ok. If the transmitter is having errors more than the maximum permissible error, then calibrate the transmitter. And after calibration, recheck all the values.
  2. Restore the original connections. Take transmitter in line.
  3. Close the back cover of the pressure transmitter and apply monsoon tape or silastic whichever is applicable.
  4. After normalizing the pressure transmitter, check the present PT values with the old values taken before the job and ask for operations that they are correct or not (sometimes due to ongoing process values may differ). If found not ok, then recheck the connections. (Optional)
  5. Apply a calibration sticker on the pressure transmitter with the appropriate calibration due date.
  6. Clean the pressure transmitter.
  7. Restore the interlocks if any and also ask the operation engineer to put the control loop again in auto mode if taken in manual mode.
  8. Close the work permit.
  9. Fill up the calibration record of the pressure transmitter.

Pressure Transmitter Checklist

The pressure transmitter PM checklist to be filled is mentioned as follows.

  1. Tag number of Instrument
  2. Order Number
  3. Permit Number
  4. Range of Instrument
  5. Date of Calibration
  6. Value before doing Job
  7. Value after doing Job
  8. Transmitter Make & Model number
  9. Test and Calibration Instrument used
  10. Test Result
  11. Technicians involved in Job
  12. Engineers involved in Job
Don't Miss Our Updates
Be the first to get exclusive content straight to your email.
We promise not to spam you. You can unsubscribe at any time.
Invalid email address
You've successfully subscribed !

Continue Reading

Pressure Measurement using Manometer
All About Instrumentation Impulse Lines
Fine Wire Strain Gauge Principle
DP Transmitter Applications
Diaphragm Seal Material Selection
DP Transmitter Block and Bleed Valves
Share This Article
Facebook Whatsapp Whatsapp LinkedIn Copy Link
Share
1 Comment
  • Takrim says:
    June 4, 2022 at 12:04 pm

    Thanks to increase in my skills

    Reply

Leave a Reply Cancel reply

Your email address will not be published. Required fields are marked *

Stay Connected

128.3kFollowersLike
69.1kFollowersFollow
210kSubscribersSubscribe
38kFollowersFollow

Categories

Explore More

Pneumatic Pressure Transmitters Principle
Sample Raw DataSheet for Pressure Calibration
Pressure Gauge Maintenance and Troubleshooting
What is Sensor Trim and Output Trim in Smart Transmitter?
Foundation Fieldbus Pressure Transmitter Configuration
Inferential Measurement based on Pressure
Identify the Pressure Gauge Error
Remote Diaphragm Seal Transmitter for Vacuum Application

Keep Learning

diaphragm-pressure-gauge

Diaphragm Pressure Sensors

pressure-switch-calibration-procedure

Pressure Switch Calibration Procedure

Pressure Gauge with Block and Bleed Valves

Pressure Gauge with Block and Bleed Valves

Semi Conductor or Piezo Resistive Strain Gauge Principle

Semi Conductor or Piezo Resistive Strain Gauge Principle

Capacitance Pressure Sensor Works

How Capacitive Pressure Sensor Works ?

Pressure Gauge Calibration according to Standard DKD-R-6-1

Pressure Gauge Calibration according to Standard DKD-R-6-1

Pressure Transmitter Circuit

Pressure Transmitter Circuit

Pressure Manometer

Types of Manometers

Learn More

Auto Transformer Schematic

Classification of Transformers

Comparison of Level Meters - Level Measurement Instruments

Comparison of Level Meters – Level Measurement Instruments

measure the temperature of RTD

Measure the Temperature of the RTD

Control Sytem Open Loop and Closed Loop

Introduction to Control System

LBB protection relay

What is the LBB Protection Relay? (Local Breaker Backup)

Stability in Frequency Domain

Time Domain Analysis

Search tank in Intouch Scada

Tank Filling and Emptying using Intouch Scada Script

Foxboro pneumatic instruments

Pneumatic Force Balance Instrument

Menu

  • About
  • Privacy Policy
  • Copyright

Quick Links

  • Learn PLC
  • Helping Hand
  • Part Time Job

YouTube Subscribe

Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Welcome Back!

Sign in to your account

Username or Email Address
Password

Lost your password?