Objective :
To check the characteristics and healthiness of vibration probes ( make : Bently Nevada )
Principle:
Proximity transducers use Eddy Current principle to measure the distance between the probe Tip and the surface to be observed.
The proximeter generate a low power radio frequency (RF) signal. This RF signal is connected to a coil of wire inside the probe tip by the extension cable.
When No conductive material is within the range of RF signal which surrounds the probe tip, virtually all of the power released to the surrounding area is returned to the probe.
When a conductive surface approaches the probe tip, the RF signal sets up small eddy currents on the surface. These eddy currents creates a measurable power loss in the RF signal.
When vibration tip nearer the target the greater the power loss. The system uses this power loss to generate as output voltage.
The output voltage of the proximeter is linearly proportional to gap over a wide range.
Required tools:
  • Multimeter
  • Micrometer
  • TK-3
Also Read : What is TK-3
Equipment Setup:
To check the characteristic of vibration probe
Procedure: Follow the following steps
1) Physically check the vibration probe and extension cable for any damages, if it is please Replace with new one.
2) Check resistance of vibration probe and continuity of extension cable it should be in between 5 To 9Ω and 5 to 20 Ω (resistance value varies from model to model)
3) Place the vibration probe on TK-3 spindle and adjust to the target plate surface and make sure the scale should be zero.
4) Connect multimeter on proximeter common and Vout terminals. Apply input voltage.
5) Measure the output voltage (multimeter reading) and increase the target distance in the TK-3 calibrator and follow the same steps as per given table and note down the output voltages at different displacements (gaps).
Bently Nevda Vibration Probes Voltage
6) After this make a graph which shows the relation between gap and voltage.
7) If graph is linear in between 10 mils to 80 mils then probe is in good condition.
Vibration Probe Calibration
Proximitor Calibration graph :
Bently Nevda Vibration Probes Calibration
Note : Change in gap is within 80 mils Linear Range which is in between 10 mils and 90 mils
Article by :
Ashish Agrawal
Displacer Level Transmitter Calibration using Water & Process Liquid
Smart Transmitter Calibration Tutorial Part 1
Internal Powered Modems
Difference between PT100, PT1000, Nickel RTD and Balco RTD
Using of MCT instead of Cable Glands on the Cabinets
H & B Gas Analyzer Principle and Calibration Procedure
9 Comments
  • rakhi rawat says:
    October 21, 2017 at 8:38 am
    its a very good n easy for learn i like too much this article
  • prasanth says:
    October 26, 2017 at 3:42 pm
    Why using negative voltage? -24vdc
    • S Bharadwaj Reddy says:
      October 26, 2017 at 3:45 pm
      Hi, check this article, CLICK HERE
  • azetou says:
    October 31, 2017 at 2:16 pm
    Thank you for the article, there is a mistake in the graph though : Gap is in mils not in millimeters.
      October 31, 2017 at 3:44 pm
      Corrected. Thanks for your efforts & updates.
  • ashraf says:
    April 15, 2018 at 11:40 pm
    Incremental Scale Factor (ISF) is just as important as Average Scale Factor (ASF). For a better proximity transducer system verification, one can easily refer to Bently Manuals. They have written the procedure in a clear and concise manner.
    Another important thing, need to confirm machinery shaft material as well whether the proximitor used is modified for linearity other than AISI 4140 (typical material for shaft). You can see physically a ‘MOD’ sticker with details on it, on the side of the proximitor itself. Performing a verification check using a MOD proximitor against a standard AISI 4140 spindle micrometer will render the result invalid.
    • RAMASAMY GANAPATHY says:
      August 8, 2023 at 2:10 pm
      In Bently Neveda Vibration Measurement, Some Equipments are having Proximitor sensor model like 330180-X1 not having scale factor 200mv/mil. Because it not Calibrated by AISI 4140 but by some other metal target place. In that case , we should not use 4140 target material disc. We have to buy corresponding material target button.
      • paco says:
        May 19, 2025 at 10:29 pm
        Hola buen dia, podras compartir mas informacion al respecto. te lo agradeceria
  • Breaker says:
    November 13, 2018 at 4:04 pm
    i couldn’t find the manual from bently nevada that talks about the procedure for checking or calibration te sensor*proximiter , could you please refere to this document or give me the link to it thanks
Magnetostrictive Level Transmitter Calibration Procedure
Ultrasonic Level Transmitter Principle, Limitations, Calibration and configuration
Seismic Transducers Objective Questions
Reversing Control Valve Actuator Fail Safe Mode
Types of Level Gauges
Determine Lifting of Shaft using Proximity Sensor
Calibration of Displacer Level Transmitter
Why You Need FDA Process Validation and How to Do It Right ?
Why we use -24vdc in vibration measurement
Why we use Negative Voltage in Vibration Measurement ?
vibration probe mounting
Mechanical Vibration Switch Principle
Difference between Calibration and Ranging
Temperature-gauge-calibration
Temperature Sensor calibration procedure
Changing Process Medium Density has no influence on Level Indication (Full Absorption)
Field Transmitter Damping
What is Field Transmitter Damping ?
differential-pressure-transmitter-calibration-procedure
Differential Pressure Transmitter Calibration Procedure
Turbidity Sensor Calibration Procedure
Control Systems Questions & Answers
Control Systems Objective Questions & Answers – Set 5
Layer-2 and Layer-3 Network Switches
Opposite Acting Valves
What is Split Range Control Loop ?
Omron PLC increment and decrement
Increment & Decrement Instructions in CX-Programmer (OMRON)
Ultrasonic level measurement
Ultrasonic Level Measurement
Supermarket HVAC Ventilation System
Supermarket Ventilation System – HVAC Basics
RDOL starter
What are DOL and RDOL starters? Advantages, Disadvantages
Top 300 Digital Signal Processing Projects (DSP)
Top 300 Digital Signal Processing Projects – DSP Project Ideas