Inst ToolsInst ToolsInst Tools
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Search
  • Books
  • Software
  • Projects
  • Process
  • Tools
  • Basics
  • Formula
  • Power Plant
  • Root Cause Analysis
  • Electrical Basics
  • Animation
  • Standards
  • 4-20 mA Course
  • Siemens PLC Course
Reading: Factors Affecting the Performance of Ultrasonic Flow Meters
Share
Notification Show More
Font ResizerAa
Inst ToolsInst Tools
Font ResizerAa
  • Courses
  • Design
  • PLC
  • Interview
  • Control System
Search
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Inst Tools > Blog > Flow Measurement > Factors Affecting the Performance of Ultrasonic Flow Meters

Factors Affecting the Performance of Ultrasonic Flow Meters

Ultrasonic Flow meter operates on principle of transit time differences. An acoustic signal is transmitted from one sensor to another to measure flow.

Last updated: November 1, 2023 6:05 am
Editorial Staff
Flow Measurement
2 Comments
Share
12 Min Read
SHARE

The ultrasonic term is used to describe pressure waves at frequencies higher than the human ears can detect.

Contents
Ultrasonic Flow MetersInstallation Options for Ultrasonic FlowmeterSingle traverse Ultrasonic Flow meterMultiple traverse Ultrasonic Flow meterFactors Affecting Performance and AccuracyA. Surface ConditionB. Transducer Positioning/AlignmentC. CalibrationD. Taper or EccentricityE. Acoustic Properties of the Material

The velocity of the sound waves in the fluid is the same as the velocity of sound in the fluid. If an ultrasonic beam is transmitted across a pipeline at an angle to the flow direction, the time taken for the pulse to reach the receiver is a function of the flow velocity of the fluid, as well as the velocity of sound in the fluid.

Thus, this type of flow meter operates on the principle of transit time differences. An acoustic signal (ultrasonic) is transmitted from one sensor to another.

Ultrasonic Flow Meters

This can be either in the direction of flow (downstream) or against the direction of flow (upstream). The time (transit) that the signal requires to arrive at the receiver is then measured.

According to physical principles, the signal sent against the direction of flow requires longer to return than the signal in the direction of flow.

The difference in the transit time is directly proportional to the velocity of flow.

v ≈ Δ t
v = k Δ t

where k is a constant

Flow rate is thus, Q = v . A

where,

v = flow velocity
Δt = transit time difference between the signal in the direction of flow and against the direction of flow
Q = volumetric flow
A = pipe cross-sectional area

Ultrasonic Flow meters Principle

Transmitter and receiver roles are constantly switched by electronics. Ultrasonic pulses traveling in the direction of the flow, travel path between transducers in a shorter period of time than pulses traveling against the flow.

This is because the flow speeds up pulses traveling downstream but slows down pulses traveling upstream. The equation is;

Ultrasonic Flow meters Formula

Installation Options for Ultrasonic Flowmeter

A tilted diameter flow cell is so named because the transducers send their pulses obliquely at a 45 angle across the diameter of the pipe. This type of flow cell can be configured as a single traverse or multiple traverse.

Single traverse Ultrasonic Flow meter

A single-traverse configuration consists of two transducers mounted across from each other on the pipe (figure 1) so that the signal they transmit passes through the fluid just once at a 45 angle.

Single traverse Ultrasonic Flow meter

Multiple traverse Ultrasonic Flow meter

A multiple-transverse configuration consists of two transducers mounted on the pipe so that the signal traverses the fluid two or more times before reaching the other transducer (figure 2). The walls of the pipe reflect the ultrasonic signal in order to maintain at 45 angle.

This allows the signal to remain in the fluid longer, increasing the effective length ( L ) of the signal path, and thus increasing accuracy.

The number of times the signal can traverse through the fluid depends on factors such as transducer frequency, pipe size, pipe wall condition, and the fluid being measured.

Multiple traverse Ultrasonic Flow meter

Four traverse Ultrasonic Flow meter

Factors Affecting Performance and Accuracy

The following factors affect the performance of ultrasonic flow meters.

  • Surface Condition,
  • Transducer Positioning/Alignment,
  • Calibration,
  • Taper or Eccentricity,
  • Acoustic Properties of the Material

A. Surface Condition

Loose or flaking scale, rust, corrosion, or dirt on the outside surface of a test piece will interfere with the coupling of sound energy from the transducer into the test material.

Thus, any loose debris of this sort should be cleaned from the specimen with a wire brush or file before measurements are attempted.

Generally, it is possible to make corrosion measurements through thin layers of rust, as long as the rust is smooth and well-bonded to the metal below.

Some very rough cast or corroded surfaces may have to be filed or sanded smooth in order to ensure proper sound coupling. It may also be necessary to remove paint if it has been applied in thick coats, or if it is flaking off the metal.

While it is often possible to make corrosion measurements through thin coats of paint (on the order of a few thousandths of an inch or 0.1 – 0.2mm), thick paint will attenuate signals or possibly create false echoes, causing inaccurate measurements.

Severe pitting on the outside surface of a pipe or tank can be a problem. On some rough surfaces, the use of a gel or grease rather than a liquid couplant will help transmit sound energy into the test piece.

In extreme cases it will be necessary to file or grind the surface sufficiently flat to permit contact with the face of the transducer.

In applications where deep pitting occurs on the outside of a pipe or tank it is usually necessary to measure remaining metal thickness from the base of the pits to the inside wall.

There are sophisticated ultrasonic techniques utilizing focused immersion transducers that can measure directly from the base of the pit to the inside wall, but this is generally not practical for field work.

The conventional technique is to measure unfitted metal thickness ultrasonically, measure pit depth mechanically, and subtract the pit depth from the measured wall thickness.

Alternately, one can file or grind the surface down to the base of the pits and measure nominally.

As with any difficult application, experimentation with actual product samples is the best way to determine the limits of a particular gage/transducer combination on a given surface.

B. Transducer Positioning/Alignment

For proper sound coupling the transducer must be pressed firmly against the test surface.

On small diameter cylindrical surfaces such as pipes, hold the transducer so that the sound barrier material visible on the probe face is aligned perpendicular to the center axis of the pipe.

While firm hand pressure on the transducer is necessary for good readings, the probe should never be scraped along or twisted against a rough metal surface. This will scratch the face of the transducer and eventually degrade performance.

The safest technique for moving a transducer along a rough surface is to pick it up and reposition it for each measurement, not to slide it along.

Remember that an ultrasonic test measures thickness at only one point within the beam of the transducer, and that in corrosion situations wall thicknesses often vary considerably.

Test procedures usually call for making a number of measurements within a defined area and establishing a minimum and/or average thickness.

Ideally, data should be taken at increments no greater than half the diameter of the transducer, to insure that no pits or other local variations in wall thickness are missed.

It is possible that on some severely corroded or pitted materials there will be spots where readings cannot be obtained.

This can happen when the inside surface of the material is so irregular that the sound energy is scattered rather than being reflected back to the transducer.

The lack of a reading may also indicate a thickness outside the range of the transducer and instrument being used.

Generally, an inability to obtain a valid thickness reading at a particular point on a test specimen could be a sign of a seriously degraded wall which may warrant investigation by other means.

C. Calibration

The accuracy of measurements are only as good as the accuracy and care with which the gauge has been calibrated.

It is essential that the velocity and zero calibrations be performed whenever the transducer is changed or you have a reason to doubt the accuracy of the readings.

Periodic checks with samples of known thicknesses are recommended to verify that the gauge is operating properly.

D. Taper or Eccentricity

If the contact surface and the back surface are tapered or eccentric with respect to each other, the return echo again becomes distorted and the accuracy of measurement is diminished.

E. Acoustic Properties of the Material

There are several conditions found in engineering materials that can severely limit the accuracy and thickness range that can be measured.

1. Sound Scattering

In some materials, notably certain types of cast stainless steel, cast irons, and composites, the sound energy is scattered from individual crystallites in the casting or from dissimilar materials within the composite.

This effect reduces the ability to discriminate a valid return echo from the back side of the material and limits the ability to gauge the material ultrasonically.

2. Velocity Variations

A number of materials exhibit significant variations in sound velocity from point-to-point within the material.

Certain types of cast stainless steels and brass exhibit this effect due to a relatively large grain size and the anisotropy of sound velocity with respect to grain orientation.

Other materials show a rapid change in sound velocity with temperature. This is characteristic of plastic materials where temperature must be controlled in order to obtain maximum precision in the measurement.

3. Sound Attenuation or Absorption

In many organic materials, such as low density plastics and rubber, sound is attenuated very rapidly at the frequencies used in normal ultrasonic thickness gauging.

Therefore, the maximum thickness that can be measured in these materials is often limited by sound attenuation.

If you liked this article, then please subscribe to our YouTube Channel for Instrumentation, Electrical, PLC, and SCADA video tutorials.

You can also follow us on Facebook and Twitter to receive daily updates.

Read Next:

  • Ultrasonic Level Switch Principle
  • Radar Level Sensor Effects
  • What is Ultrasonic Testing (UT)
  • Transit Time Flow Meter Uses
  • Vortex Flow Meter Sensors
Don't Miss Our Updates
Be the first to get exclusive content straight to your email.
We promise not to spam you. You can unsubscribe at any time.
Invalid email address
You've successfully subscribed !

Continue Reading

Basics of Flow Measurement
Venturi Flow Meter – Working Principle, Advantages, Animation
What is an Orifice Meter?
Why Turndown Ratio is important when Selecting a Flow Meter?
Orifice Sizing Guidelines and Thumb Rules with Flowchart
Microwave Flow Sensor – Working Principle and Applications
Share This Article
Facebook Whatsapp Whatsapp LinkedIn Copy Link
Share
2 Comments
  • Yuriy says:
    July 21, 2018 at 7:28 pm

    Hi! Thank’s for this post!I have one question-what is a K (meter factor)?How can i calculate it?

    Reply
    • S Bharadwaj Reddy says:
      July 22, 2018 at 12:22 pm

      Hi, Please check “Flow meter K factor” article. CLICK HERE

      Reply

Leave a Reply Cancel reply

Your email address will not be published. Required fields are marked *

Stay Connected

128.3kFollowersLike
69.1kFollowersFollow
210kSubscribersSubscribe
38kFollowersFollow

Categories

Explore More

What are Multiphase Flow Meters? – Working Principle, Advantages
Orifice Plate Tapping – Orifice Plate Taps
80+ Flow Measurement Interview Questions and Answers
Difference between Fixed Area & Variable Area FlowMeters
Oval Gear Meter – Principle, Advantages, Limitations, Applications
Weirs and Flumes Flow Measurement
Classification of Flow Measurement Techniques
DP Flow Transmitter Re-Ranging Calculation

Keep Learning

Orifice Flange Rating

Why Orifice Flow Meter Flange is Not Recommended 150# Rating?

Flow Switch Working Principle Animation

Working Principle of Flow Switch with Animation

Spring Loaded Variable Area Flowmeter Working Principle

Spring Loaded Variable Area Flow Meter Working Principle

Flow Meter Calibration Procedure

Flow Meter Calibration

Ultrasonic Flow meters Working Animation

Ultrasonic Flow Meters Animation

Annubar Flow Meter Principle

Annubar Flow Meter – Principle, Advantages, Disadvantages

Orifice Specifications

Orifice Flow Meters: Detailed Specifications

Variable Area Flow Meters

All About Variable Area Flow Meters

Learn More

Allen-Bradley MicroLogix PLC Wiring

Programmable Logic Controller (PLC) Questions and Answers – 8

Electrical Machines Questions and Answers

Starting of Polyphase Induction Motors Quiz

OB10 Time of Day Interrupt

Tia Portal – OB10 Time of Day Interrupt Organization Block

Selector Control Function in DCS

Selector Control Functions

Differential Pressure Transmitter Wet Leg Compensation

Differential Pressure Transmitters Compensated Leg Systems

Capacitor Bank

What is a Capacitor Bank?

What is Partial Stroke Test

What is Partial Stroke Test?

Half-controlled Rectifier With R Load

Phase Controlled Rectifier – Half Controlled Rectifier with R Load

Menu

  • About
  • Privacy Policy
  • Copyright

Quick Links

  • Learn PLC
  • Helping Hand
  • Part Time Job

YouTube Subscribe

Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Welcome Back!

Sign in to your account

Username or Email Address
Password

Lost your password?