Inst ToolsInst ToolsInst Tools
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Search
  • Books
  • Software
  • Projects
  • Process
  • Tools
  • Basics
  • Formula
  • Power Plant
  • Root Cause Analysis
  • Electrical Basics
  • Animation
  • Standards
  • 4-20 mA Course
  • Siemens PLC Course
Reading: What is a Dry Block? – Parts of the Temperature Calibrator
Share
Font ResizerAa
Inst ToolsInst Tools
Font ResizerAa
  • Courses
  • Design
  • PLC
  • Interview
  • Control System
Search
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Inst Tools > Blog > Calibration > What is a Dry Block? – Parts of the Temperature Calibrator

What is a Dry Block? – Parts of the Temperature Calibrator

Dry Block is used to heat or cool to various temperature ranges in order to calibrate the temperature sensors in a calibrator.

Last updated: March 10, 2023 2:16 pm
Shiv Kumar
Calibration
1 Comment
Share
8 Min Read
SHARE

What is a Dry Block?

  • Dry Block is a kind of device that can be heated or cooled to various temperature ranges.
  • The name Dry Block suggests that it is used dry, without using any liquid materials.
  • A temperature dry block is known as a temperature calibrator sometimes it is called a dry well.
  • Generally, a Dry block consists of a removable sleeve or insert with suitable holes or bores for inserting appropriate sensors into the dry block.
  • The dry block consists of an internal temperature measurement arrangement,
  • Sometimes we may use an external reference temperature sensor to insert into holes on the dry block.
  • Commonly a dry block temperature calibrator consists of several compatible inserts.
  • Each insert is drilled with bores of different sizes to match for calibration of varying-sized temperature sensors or transducers.
  • The temperature sensor bores in the dry block are tight enough to enable low thermal resistance between the sensor and the insert provided.
  • If the bores provided are too loose then the sensor gets stabilized slowly, but sometimes they are unable to reach the insert temperature due to stem conditions.
  • Generally, when we insert a temperature sensor or transducer to be calibrated in a dry block, the temperature sensor is the principal component to calibrate a temperature loop.
  • All dry block temperature calibrators have routes to
  • Sourcing Temperature
  • Reading, Analyzing, & Reporting the Temperature.

Sourcing Temperature

Two common methods of sourcing temperature calibrators include

Contents
What is a Dry Block?Sourcing TemperatureA. Simulation B. Physical Heat GenerationReading, Analyzing & Reporting TemperatureFeatures of Dry Well CalibratorParts of the Dry Block CalibratorSpecification of Dry Block Temperature CalibratorWhat are the other names of a Dry Block temperature Calibrator?Advantages of a Dry Block CalibratorDisadvantages of a Dry Block Calibrator

A. Simulation 

  • Temperature calibrators using simulation hand-held devices.
  • Its internal electronic circuitry produces
  • Different voltage levels are usually in terms of millivolts.
  • Resistance to simulate temperature sensors.
  • An output voltage signal generated represents a thermocouple unit.
  • An output resistance signal represents RTD, Thermistor, or PRT.
  • A simulator verifies sensor readouts such as for temperature transmitters or oven displays.
  • This simulator measures and displays temperature sensors’ electrical output correctly.
  • The simulator allows the user to select a temperature sensor type to simulate during operation.

B. Physical Heat Generation

  • Temperature calibrator generates physical heat including
  • Dry-Block Furnaces,
  • Infrared Radiation (Ir) Calibrators,
  • Temperature Chambers,
  • Ovens & Temperature Calibration Baths.
  • This calibrator consists of built-in sensors for measuring and displaying the calibrator temperature.
  • These are considered and used as reference sensors then built-in sensors having low accuracy.
  • It is required to insert sensors having high accuracy called external reference probes such as Platinum Resistance Thermometers, or Resistance Temperature Devices to obtain more accuracy and low uncertainty.

These are inserted into the heat source along with the device under test (DUT).

Reading, Analyzing & Reporting Temperature

  • There are some common methods by that a temperature calibrator can read, analyze and report temperatures.
  • The common way for the device to source the temperature must have an input to measure the DUT sensor.
  • This Dry block temperature calibrator displays the temperature of the dry block and Device under test.
  • Another common method is inserting a sensor into the calibrator, and the sensor stays connected to the display, & transmitter.
  • This method displays both temperatures of the calibrator & DUT.
  • This method is recommended, & considered as best because it compares and verifies both temperature of the calibrator, & the DUT sensor.
  • Some temperature calibrators are built with an interface unit for the user to link a reference digital thermometer system to determine the highest accuracy

Features of Dry Well Calibrator

  • Temperature ranges from -40℃ to 2200℃
  • It has a Portable design & rugged structure.
  • Accuracy is good
  • Rapid to-temperature climatic science performance in their stability
  • Dual-zone control.
  • For reference Thermometer, RTDs, and T/Cs dry block calibrator provides a multi-channel readout for use.
  • The dry block calibrator provides task documentation, switch testing, and HART communication.
  • Touch screen with Color Display interface.
  • Provides an option to choose your own range.
  • Set point control by reference PRT
  • Self-calibration feature 

Parts of the Dry Block Calibrator

Parts of the Temperature Calibrator

The dry Block Calibrator consists of

  • A well-designed insert hole for lodging various devices under test or sensors to be calibrated.
  • Heating coils for the escalating temperature to the required value.
  • Reference thermometer fixed internally within dry block calibrator to determine and display device temperature.
  • An electronic PID controller is required to provide a temperature Setpoint and to control the temperature of the Dry Block calibrator.

Specification of Dry Block Temperature Calibrator

Range300°C to 1200°C
Resolution1°C
Accuracy±2°C
Stability±0.5°C
Stabilization Time15 Minutes
Temperature Readout°C / °F Selectable
Well Diameter26 mm
Immersion Depth105 mm
Dry Block MaterialInconel
Heating period60 Minutes approx.
Cooling Time (to min.)45 Minutes approx. (with Forced Air Cooling)
Switch Test facilityProvided
Thermowell Diameter8 mm
Housing / MountingBench-top
Power Supply230V AC, 50 Hz
Instrument Dimensions160 x 365 x 450 mm  (L x D x H)
Instrument Weight15 Kg.
Measuring CapacityThermocouples J, K, R, S, N, B, T
RTDPt100 IEC Std.
mA0 to 25 mA
mV0 to 100 / 1000 mV
Volt0 to 10 V
Resistance0 to 500Ω / 3.5 KΩ

What are the other names of a Dry Block temperature Calibrator?

Commonly these are called by

  • Dry Block
  • Dry Block Temperature Calibrator
  • Dry-Well Calibrator
  • Block Calibrator
  • Temperature Block
  • Dry Bath Temperature Calibrator

Advantages of a Dry Block Calibrator

  • Dry block calibrators are easy to carry out in the field.
  • It does not contain any hot fluid that may spill during carrying.
  • These devices do not contaminate the calibrating temperate sensors.

Disadvantages of a Dry Block Calibrator

  • Accuracy is less.
  • Compared to a liquid bath it offers less stability.
  • Calibration is difficult for sensors having short stems and oddly shaped ones.
  • Not applicable for liquid use.
  • Requires some heat transfer fluids or pastes in rare cases.

If you liked this article, then please subscribe to our YouTube Channel for Instrumentation, Electrical, PLC, and SCADA video tutorials.

You can also follow us on Facebook and Twitter to receive daily updates.

Read Next:

  • Calibration Organizations
  • Create Calibration Records
  • Methods of Measurement
  • Principle of Dead Weight
  • Uncertainty of Calibration
Don't Miss Our Updates
Be the first to get exclusive content straight to your email.
We promise not to spam you. You can unsubscribe at any time.
Invalid email address
You've successfully subscribed !

Continue Reading

Automated Calibration
Pressure Gauge Functional Testing
Hybrid Level Measurement Principle, Selection, Installation, Limitations and Calibration
RTD Calibration
Common Mistakes to Avoid When Using Dead Weight Tester
Pressure Gauge Calibration Procedure
Share This Article
Facebook Whatsapp Whatsapp LinkedIn Copy Link
Share
1 Comment
  • Raju Yadav says:
    March 31, 2023 at 1:48 pm

    Please send heat treatment furnace calibration procedures

    Reply

Leave a Reply Cancel reply

Your email address will not be published. Required fields are marked *

Stay Connected

128.3kFollowersLike
69.1kFollowersFollow
210kSubscribersSubscribe
38kFollowersFollow

Categories

Explore More

Instrument Turndown Ratio
Why Calibration is Important?
Calibration Procedure of Weighing Balance and Common Mistakes
What is Sensor Trim and Output Trim in Smart Transmitter?
Selection of Industrial Temperature Calibrator
What is Bump Testing ?
Calibration of Measuring Instruments – Significance, Costs & Risks
Current to Pressure (I/P) Converter Calibration Procedure

Keep Learning

Closed-tank-DP-Level-Transmitter-with-cleaning-leg-Calibration

Closed tank DP Level Transmitter with cleaning leg Calibration

Calibration Procedure of displacer type Level transmitter

Calibration of Displacer Level Transmitter

Calibrate Eddy Current Conductivity Instrument

How to Calibrate the Eddy Current Conductivity Meter?

Calibration of Weight

Calibration of Weight using the ABBA Method

Temperature Switch Calibration

How to Calibrate Temperature Switch? – Instrument Basics

DP type Flow Transmitter Calibration

DP type Flow Transmitter Preventive Maintenance and Calibration

DP-Level-Transmitter-Open-tank-calibration

Open Tank DP Level Transmitter Calibration

Error sources creating Uncertainty in calibration

Error sources creating uncertainty in Calibration

Learn More

Main Functions of a Multimeter

Interesting Multimeter Functions (And What They Really Do)

Transistor Emitter Feedback Bias

Transistor Emitter Feedback Bias

Electrical Machines Questions and Answers

DC Machines Excitation Quiz

AND Gate Logic Animation

Digital Building Blocks – GATES

Questions and Answers on Valves

Valve Quiz

DTS sensors

Distributed Temperature Sensors (DTS) in Power Lines

Types of Engineering Courses

Various Types of Engineering Courses

Formula for 4-20mA from 3-15PSI

Formula to Calculate 3-15psi from 4-20mA Current

Menu

  • About
  • Privacy Policy
  • Copyright

Quick Links

  • Learn PLC
  • Helping Hand
  • Part Time Job

YouTube Subscribe

Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Welcome Back!

Sign in to your account

Username or Email Address
Password

Lost your password?