Inst ToolsInst ToolsInst Tools
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Search
  • Books
  • Software
  • Projects
  • Process
  • Tools
  • Basics
  • Formula
  • Power Plant
  • Root Cause Analysis
  • Electrical Basics
  • Animation
  • Standards
  • 4-20 mA Course
  • Siemens PLC Course
Reading: H & B Gas Analyzer Principle and Calibration Procedure
Share
Notification Show More
Font ResizerAa
Inst ToolsInst Tools
Font ResizerAa
  • Courses
  • Design
  • PLC
  • Interview
  • Control System
Search
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Inst Tools > Blog > Analyzers > H & B Gas Analyzer Principle and Calibration Procedure

H & B Gas Analyzer Principle and Calibration Procedure

Last updated: December 20, 2019 3:49 pm
Editorial Staff
Analyzers Calibration Guest Articles
2 Comments
Share
6 Min Read
SHARE

H & B Gas Analyzer Principle and Calibration Procedure

Objective:

H& B make gas analyzer (For Measuring CH4, CO2, and Co) Principle & Calibration Procedure.

Principle:

Gas analyzers works on NDIR (Non dispersive Infrared sensor)

The main components of NDIR sensors are an infrared source, a sample chamber and an infrared detector. The IR light is passed through the sample cell & Reference cell and reaches the detector. Reference cell is filled with a gas, typically nitrogen.

gas Analyzer

The gas to be measured is called as sample gas. This sample gas passes through the sample cell. A chopper is used to split the source i.e. IR rays to pass via both sample and reference cells.

The gas in the sample cell causes absorption of specific wavelengths according to the Beer–Lambert law, and the attenuation of these wavelengths is measured by the detector to determine the gas concentration. The detector has an optical filter in front of it that eliminates all light except the wavelength that the selected gas molecules can absorb.

Also See : NDIR Gas Analyzer Animation

The relationship between measured infrared radiation absorption and the measurement component is based on the LAMBERT-BEER law:

Gas Analyzer Formula

Where

A = Absorption

I0 = Radiation entering the cell

I1 = Radiation leaving the cell

ε(λ) = Sample component extinction factor

ρ = Sample component density

l = Sample cell length

Infrared rays: Infrared rays is a part of electromagnetic wave of which wavelength is longer than visible red.

Infrared Wavelength Range

Parts of Gas analyzers:

H & B Gas Analyzer Display unit

Display Unit: Display unit shows the concentration of gas values and for user interface.

IR radiation: Generated by broad-band emitter

Sample Cell: In sample cell there are two chambers 1. Reference chamber 2. Sample gas chamber. Reference chamber is filled with Inert gas like as N2 gas. Cell length differs according to the measurement range. The IR radiations will be spitted by chopper motor and passes through both sample and reference chambers and reaches the detector.

Gas Detector: The detector measures the intensity of IR rays which are received after passing via sample and reference cells. The IR rays intensity is measured and converts into equivalent gas concentration readings.

The preamplifier has 3 jumpers to allow signal matching to x1, x3 or x10. These jumpers are set during the initial calibration

Chopper Motor: Instead of using two individual sources or IR rays transmitters, we can use single source with a Chopper motor. The chopper splits the IR rays from the source to pass through the reference and sample cells. So that the infrared light from the source/emitters passes alternately through each cell twice per revolution.

H & B Gas Analyzer

Configuration of analyzers:

For configuration of H&B URAS 14 : install the TCT software and communicate with analyzer then open its software application, following page opened

configuration H&B URAS 14

  1. Add the URAS 14 detector 1 and select the gas which you want to measure and which detector is installed. Then enter the gas concentration range.
  2. Add a pressure detector and enter the range ( say 0-100 kpa )
  3. Add a temperature detector and enter range (say  0 – 100 degc )

After this send the data to sample cell unit.

Calibration Procedure:

Make: H&B,  Model: Uras 14

For calibration,  go to calibration mode by display menu then enter into manual calibration menu.

  1. Isolate the sample gas and open the valve of zero gas (N2 gas) , then select zero calibration menu then enter the concentration of zero gas value, press enter and wait for sometime to stable the readings, after stabilizing the reading press enter for zero calibration.
  2. For span gas calibration open the valve of span gas, then select span gas calibration menu then enter the value of span gas cylinder concentration value and wait for sometime to stable the readings, after stabilizing the reading press enter for Span calibration

After this go to measuring display menu.

Sampling System :

For any gas analysis, to obtain respective sample of the stream is very important. The sampling system equipment’s should be able to continuously purify and deliver the gas sample to the gas analyzer.

If sample gas has dust, condensate and corrosive substances then these may disturb the analyzer performance and its analysis, therefore sampling system should include dust filters, measuring gas cooler, absorption filter, mist filters and condensate separators etc.

The sample gas flow and the bypass flow must be connected to the atmosphere or to an unpressurized exhaust gas collecting system at an adequate distance from the Gas analyzer.

Routine Activity for Gas analyzers :

  • Sample condensate flushing /drain moisture catch pot from its primary sampling system as well secondary sampling system
  • Sample pressure to be noted and maintained at required value
  • Ensure proper Sample flow, Pressure & temperature values
  • Monitor electronics/Check for any alarm messages and take care timely action.
  • Timely Calibrate the gas analyzers as PM schedule.
Article by :
Ashish Agrawal
Don't Miss Our Updates
Be the first to get exclusive content straight to your email.
We promise not to spam you. You can unsubscribe at any time.
Invalid email address
You've successfully subscribed !

Continue Reading

Single-beam Non-dispersive Analyzer
PLC Program for Automatic Liquid Mixing Application
Role of Flow Meters in a Laboratory
Scintillation Counter Principle
Ionization chamber Principle
Conductivity Analyzers Interview Questions & Answers
Share This Article
Facebook Whatsapp Whatsapp LinkedIn Copy Link
Share
2 Comments
  • Harikumar says:
    November 28, 2017 at 9:53 am

    Plz ABB GAS ANALYSER DETAILS

    Reply
  • Steve perry says:
    August 18, 2021 at 11:00 am

    Hi.

    Do you know where I can train to calibrate gas flue analysers in the UK?

    Many thanks, Steve

    Reply

Leave a Reply Cancel reply

Your email address will not be published. Required fields are marked *

Stay Connected

128.3kFollowersLike
69.1kFollowersFollow
210kSubscribersSubscribe
38kFollowersFollow

Categories

Explore More

Hierarchy of Calibration Organizations
Humidity Measurement Principle
Installation and Calibration of Pressure Transmitter
Gas Analyzers Interview Questions and Answers
Non-Dispersive Analyzers
PLC Program for Automatic Bottle Rejection System
Difference between Calibration and Ranging
PLC Program for Positive Edge Pulse Output for One Scan Cycle

Keep Learning

List of Documents Required for Testing and Calibration Laboratory

List of Documents Required for Testing and Calibration Laboratories

Optical Pyrometer

Optical Pyrometer Working Principle

As-Found and As-Left data during Calibration

Why it is important to note As-Found and As-Left data during Calibration?

Gas Filter Correlation (GFC) spectrometer

Gas Filter Correlation (GFC) Spectroscopy

In-Situ Zirconia Oxygen Analyzer

Oxygen (O2) Measurement in Flue Gas

Calibrate Eddy Current Conductivity Instrument

How to Calibrate the Eddy Current Conductivity Meter?

Displacer Level Transmitter weight calibration

Displacer Level Transmitter Dry Calibration with Weights

Stroke Checking Procedure for GCV, SRV, IGV, and LFBV

Stroke Checking Procedure for GCV, SRV, IGV, and LFBV

Learn More

Industrial Light Fittings and Street Light Poles Corrosion Problems

Industrial Light Fittings and Street Light Poles Corrosion Problems

Calibration of a Strain Gage Load Cell

Shunt Calibration of a Strain Gauge Load Cell

PLC Ladder Logic Door Open and Closing System

Which Language is Best for PLC Programming?

What is Fail-safe in Siemens PLC?

What is Fail-safe in Siemens PLC?

Types of PLC Memory

Types of PLC Memory

Capacitance-Level-Transmitter

Capacitance Level Switch Calibration

What is a Network

What is a Network ?

Color detection PLC logic

Conveyor Sorting System with Color Detection PLC Logic

Menu

  • About
  • Privacy Policy
  • Copyright

Quick Links

  • Learn PLC
  • Helping Hand
  • Part Time Job

YouTube Subscribe

Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Welcome Back!

Sign in to your account

Username or Email Address
Password

Lost your password?