Reading: Temperature Coefficient of Resistance (α)
Inst Tools > Blog > Temperature Measurement > Temperature Coefficient of Resistance (α)
Last updated: May 12, 2019 5:56 pm
A Resistive Temperature Detector (RTD) is a special temperature-sensing element made of fine metal wire, the electrical resistance of which changes with temperature as approximated by the following formula:
Where,
RT = Resistance of RTD at given temperature T (ohms)
Rref = Resistance of RTD at the reference temperature Tref (ohms)
α = Temperature coefficient of resistance (ohms per ohm/degree)
The following example shows how to use this formula to calculate the resistance of a “100 ohm” platinum RTD with a temperature coefficient value of 0.00392 at a temperature of 35 degrees Celsius:
RT = 100 [1 + (0.00392)(35o C − 0o C)]
RT = 100 [1 + 0.1372]
RT = 100 [1.1372]
RT = 113.72
Due to nonlinearities in the RTD’s behavior, this linear RTD formula is only an approximation. A better approximation is the Callendar-van Dusen formula, which introduces second, third, and fourth-degree terms for a better fit:
RT = Rref (1 + AT + BT2 − 100CT3 + CT4) for temperatures ranging −200 oC < T < 0 oC and
RT = Rref (1+AT +BT2) for temperatures ranging 0 oC < T < 661 oC,
both assuming Tref = 0 oC. The A, B, and C coefficients vary with the specific type of RTD, equivalent in role to α in the linear RTD formula.
Water’s melting/freezing point is the standard reference temperature for most RTDs. Here are some typical values of α for common metals:
  • Nickel = 0.00672  Ω/ΩoC
  • Tungsten = 0.0045  Ω/ΩoC
  • Silver = 0.0041  Ω/ΩoC
  • Gold = 0.0040  Ω/ΩoC
  • Platinum = 0.00392  Ω/ΩoC
  • Copper = 0.0038  Ω/ΩoC
As mentioned previously, platinum is a common wire material for industrial RTD construction. The alpha (α) value for platinum varies according to the alloying of the metal. For “reference grade” platinum wire, the most common alpha value is 0.003902. Industrial-grade platinum alloy RTD wire is commonly available in two different coefficient values: 0.00385 (the “European” alpha value) and 0.00392 (the “American” alpha value), of which the “European” value of 0.00385 is most commonly used (even in the United States!).
An alternative to mathematically predicting the resistance of an RTD is to simply look up its predicted resistance versus temperature in a table of values published for that RTD type. Tables capture the nuances of an RTD’s non-linearity without adding any mathematical complexity: simply look up the resistance corresponding to a given temperature, or vice-versa. If a value falls between two nearest entries in the table, you may interpolate the find the desired value, regarding the two nearest table entries as end-points defining a line segment, calculating the point you desire along that line.
100 Ω is a very common reference resistance (Rref at 0 degrees Celsius) for industrial RTDs. 1000 Ω is another common reference resistance, and some industrial RTDs have reference resistances as low as 10 Ω . Compared to thermistors with their tens or even hundreds of thousands of ohms’ resistance, an RTD’s resistance is comparatively small. This can cause problems with measurement, since the wires connecting an RTD to its ohmmeter possess their own resistance, which will be a more substantial percentage of the total circuit resistance than for a thermistor.
What is a Probe Thermometer? How to Calibrate and Use it?
Multiple Position Calibration of Temperature Chamber
Thermocouple Calibration
How to Select Right Thermocouple & Thermowell ?
What is Thermowell and How-to choose Thermowell Probe Length ?
Relationship between Temperature Scales
What to Choose RTD or Thermocouple ?
Applications of Temperature Scanner
RTD Wheatstone Bridge Circuits
Thermocouple Construction
Mechanical Temperature Measuring Sensors
What is a Four-wire RTD ?
What is a Thermocouple ?
Calibration of Pyrometers using Black Body
Thermocouple Law of Intermediate Metals
Thermocouples Law of Intermediate Metals
RTD Calibration
RTD Calibration
Thermocouple Calculations
Thermowell Problems
Thermowell Problems
Temperature Switch
Temperature Switch Working Principle
Measuring temperature with RTD
How RTD Measuring the Temperature
RTD-standards
How to calculate the tolerance of a RTD sensor
Stability in Frequency Domain
electric motor actuator principle
What is a Electric Motor Actuator ?
Coriolis Mass Flow meter as a Multi Variable Transmitter - 2
Coriolis Flowmeter Advantages and Limitations
Difference Between Alternator and Generator
Difference Between Alternator and Generator?
Flow Measurement Questions and Answers
Difference Between MPCB and MCCB
Piston No-Diaphragm Seal Instrument
Seal Pressure Instruments Root Cause Analysis
Half-Wave Rectifier Circuit