Inst ToolsInst ToolsInst Tools
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
Search
  • Books
  • Software
  • Projects
  • Process
  • Tools
  • Basics
  • Formula
  • Power Plant
  • Root Cause Analysis
  • Electrical Basics
  • Animation
  • Standards
  • 4-20 mA Course
  • Siemens PLC Course
Reading: Single Position Calibration of Temperature Bath
Share
Notification Show More
Font ResizerAa
Inst ToolsInst Tools
Font ResizerAa
  • Courses
  • Design
  • PLC
  • Interview
  • Control System
Search
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Inst Tools > Blog > Calibration > Single Position Calibration of Temperature Bath

Single Position Calibration of Temperature Bath

This article covers single position calibration of temperature sources. Here the unit under test (UUC) is the temperature bath.

Last updated: September 28, 2022 3:24 pm
Saurabh Vaidya
Calibration Temperature Measurement
No Comments
Share
3 Min Read
SHARE

This article covers Single Position Calibration of Temperature Bath.

Temperature Baths are available in many types.

They are as follows.

  • Liquid bath,
  • Dry well block,
  • Oven, Incubator,
  • Muffle Furnace,
  • Refrigerator,
  • Freezer, and
  • Temperature Chamber.

The temperature calibrators are also needed to be calibrated periodically. Here the unit under test (UUC) is the temperature bath.

Calibration Procedure of Temperature Bath – Single Position

The following calibration procedure will help you to calibrate the temperature bath with a single position method.

Single Position Calibration of Temperature Bath

Step 1:

Insert the Master RTD Sensor/Thermocouple into the Temperature Bath (UUC) at Center Position of Temperature Bath.

Now connect a calibrated multi-function calibrator to this master sensor to read the sensor output in temperature units like degree C.

If you don’t have a multi-function calibrator then you can use calibrated Multimeter. Note down the output and do the conversion to temperature using RTD or thermocouple standard temperature tables.

Step 2:

Decide the calibration points of the UUC before starting the actual calibration.

For e.xample

  • The calibration points of the Oven are 50, 100,150, 200 degrees Celsius.
  • The Calibration points of the Refrigerator are 2, 5 10, 15 degrees Celsius.

Discuss Calibration points with the end-user and according to the working range of the instrument, calibration points are decided.

If you had an old calibration record file then you can find the respective calibration points details in it.

Step 3:

Set the first calibration point on Temperature Bath (UUC).

Allow temperature bath to attend the desired set temperature and Stabilize at first calibration point.

Step 4:

Note down the UUC temperature reading displayed on the indicator of Temperature Bath and Master RTD Sensor/Thermocouple reading.

The temperature Bath shows two readings, Set temperature, and Actual Temperature. Note down the Actual temperature reading.

Take 5 readings at a single different setpoint in intervals of 30 sec.

Calculate the average reading for each calibration point.

Step 5:

Repeat Step 3 and Step 4 for other calibration points.

i.e., do the same calibration step for 100, 150, and 200 degrees Celsius.

Record all the readings in your calibration sheet.

If you liked this article, then please subscribe to our YouTube Channel for Instrumentation, Electrical, PLC, and SCADA video tutorials.

You can also follow us on Facebook and Twitter to receive daily updates.

Read Next:

  • RTD Calibration Procedure
  • Selection of Master Instrument
  • Importance of Unit Measurement
  • Why Calibration is Important?
  • Calibration of Level Transmitter
Don't Miss Our Updates
Be the first to get exclusive content straight to your email.
We promise not to spam you. You can unsubscribe at any time.
Invalid email address
You've successfully subscribed !

Continue Reading

Skin Type Thermocouple
Magnetostrictive Level Transmitter Calibration Procedure
What is a Head Mount Temperature Transmitter? – Advantages
Thermistor Working Principle
List of Documents Required for Testing and Calibration Laboratories
Helix Bimetallic Thermometer Principle
Share This Article
Facebook Whatsapp Whatsapp LinkedIn Copy Link
Share
Leave a Comment

Leave a Reply Cancel reply

Your email address will not be published. Required fields are marked *

Stay Connected

128.3kFollowersLike
69.1kFollowersFollow
208kSubscribersSubscribe
38kFollowersFollow

Categories

Explore More

Temperature Gauges and Elements : Detailed Specifications
Resistance Temperature Detectors Troubleshooting Tips
Selection of Master Instrument for Calibration
Calibration Procedure of Weighing Balance and Common Mistakes
What to Choose RTD or Thermocouple ?
Calculate Temperature Coefficient of RTD
Calibration MCQ – Terminology and Glossary Questions
Automated Calibration

Keep Learning

Hybrid Level Transmitter Principle

Hybrid Level Measurement Principle, Selection, Installation, Limitations and Calibration

RTD Calibration

RTD Calibration

Thermowell Immersion Length

Thermowell Insertion and Immersion Length

Closed-tank-DP-Level-Transmitter-with-cleaning-leg-Calibration

Closed tank DP Level Transmitter with cleaning leg Calibration

Thermocouple Calibration

Thermocouple Calibration

Vibration Probe Field Setting Procedure

How to do calibration checks of vibration Probe, extension cable and vibration monitor

DP type Flow Transmitter Calibration

DP type Flow Transmitter Preventive Maintenance and Calibration

As-Found and As-Left data during Calibration

Why it is important to note As-Found and As-Left data during Calibration?

Learn More

Which instrument is faulty

Which Instrument is Faulty? Transmitter, Recorder, or Controller

Digital Control Systems

Difference Equations

Digital Electronics Multiple Choice Questions

Encoders Objective Questions

Why is White Powder Used in Electrical Cables

Why is White Powder Used in Electrical Cables?

Piping and Instrumentation Diagram

What is Piping and Instrumentation Diagram (P&ID) ?

Digital Electronics MCQ

Digital Integrated Circuits Objective Questions – Part 1

Variable Frequency Drive - Converter

Variable Frequency Drive (VFD) Working Principle

Vibration Monitoring System Troubleshooting

Vibration Monitoring System Step by Step Troubleshooting Guide

Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Welcome Back!

Sign in to your account

Username or Email Address
Password

Lost your password?