Reading: Single Position Calibration of Temperature Bath
Inst Tools > Blog > Calibration > Single Position Calibration of Temperature Bath
This article covers single position calibration of temperature sources. Here the unit under test (UUC) is the temperature bath.
Last updated: September 28, 2022 3:24 pm
Saurabh Vaidya
Calibration Temperature Measurement
3 Min Read
This article covers Single Position Calibration of Temperature Bath.
Temperature Baths are available in many types.
They are as follows.
  • Liquid bath,
  • Dry well block,
  • Oven, Incubator,
  • Muffle Furnace,
  • Refrigerator,
  • Freezer, and
  • Temperature Chamber.
The temperature calibrators are also needed to be calibrated periodically. Here the unit under test (UUC) is the temperature bath.
Calibration Procedure of Temperature Bath – Single Position
The following calibration procedure will help you to calibrate the temperature bath with a single position method.
Step 1:
Insert the Master RTD Sensor/Thermocouple into the Temperature Bath (UUC) at Center Position of Temperature Bath.
Now connect a calibrated multi-function calibrator to this master sensor to read the sensor output in temperature units like degree C.
If you don’t have a multi-function calibrator then you can use calibrated Multimeter. Note down the output and do the conversion to temperature using RTD or thermocouple standard temperature tables.
Step 2:
Decide the calibration points of the UUC before starting the actual calibration.
For e.xample
  • The calibration points of the Oven are 50, 100,150, 200 degrees Celsius.
  • The Calibration points of the Refrigerator are 2, 5 10, 15 degrees Celsius.
Discuss Calibration points with the end-user and according to the working range of the instrument, calibration points are decided.
If you had an old calibration record file then you can find the respective calibration points details in it.
Step 3:
Set the first calibration point on Temperature Bath (UUC).
Allow temperature bath to attend the desired set temperature and Stabilize at first calibration point.
Step 4:
Note down the UUC temperature reading displayed on the indicator of Temperature Bath and Master RTD Sensor/Thermocouple reading.
The temperature Bath shows two readings, Set temperature, and Actual Temperature. Note down the Actual temperature reading.
Take 5 readings at a single different setpoint in intervals of 30 sec.
Calculate the average reading for each calibration point.
Step 5:
Repeat Step 3 and Step 4 for other calibration points.
i.e., do the same calibration step for 100, 150, and 200 degrees Celsius.
Record all the readings in your calibration sheet.
  • RTD Calibration Procedure
  • Selection of Master Instrument
  • Importance of Unit Measurement
  • Why Calibration is Important?
  • Calibration of Level Transmitter
Stroke Checking Procedure for GCV, SRV, IGV, and LFBV
Why 4 wire RTD Measurement Accuracy is better than 2 and 3 wire RTD?
How-to do Control Valve Calibration?
Thermocouple Questions and Answers
Effect of Lead Wire Resistance on RTD
Instruments Calibration Procedures
Radiation Pyrometer Working Principle
Calculate Temperature Coefficient of RTD
Thermocouple Temperature Measurement Errors
Pressure Gauge Calibration Procedure
Thermistor Working Principle
RTD Calculator Software
Bimetallic Thermometer
Instrument Zero and Span Calibration
Difference between Accuracy, Tolerance, Uncertainty, and Error
Pressure Gauge Calibration according to Standard DKD-R-6-1
Ultrasonic Level Transmitters Installation
Ultrasonic Level Transmitter Principle, Limitations, Calibration and configuration
Four Wire RTD Principle
How a 4 Wire RTD Works ?
Bimetallic strip thermostat
Mechanical Temperature Measuring Sensors
RTD Accuracy
What is RTD Accuracy ?
thermocouple types
What is a Thermocouple ?
process chemical analyzer calibration
Automated Calibration
Online time of the PLC
System and Local Time in PLC
Cisco CCNP Certification
Cisco CCNP R&S : All Actual Information About This Certification
Electrical Motor Failures
Most General Reasons for the Electrical Motor Failures
SCADA and PLC Configuration
Examples of SCADA and PLC Configuration Systems
Selecting the Proper Failure Mode
How to Select Control Valve Failure Mode
OPC Server
What is the OPC Server?
Emergency ShutDown Valve Working Principle
ESDV : How it Works ?
DC Motor Action
DC Motor Speed