Reading: pH sensor calibration procedure
Inst Tools > Blog > Calibration > pH sensor calibration procedure
Last updated: April 25, 2017 11:07 am
Calibration
3 Min Read
A pH meter calibration is performed by measuring a series of reference standards, so-called pH buffers that have known and accurate pH values at different temperatures.
Choose the appropriate buffered standards that will bracket the expected values at the sampling locations. If the water body’s pH is unknown, then three standards are needed for the calibration: one close to seven, one at least two pH units below seven, and the other at least two pH units above seven. Instruments that will not accept three standards will need to be re-calibrated if the sample’s pH is outside the initial calibration range.
pH Sensor calibration procedure
Calibration Procedure:
1. Allow the buffered standards to equilibrate to the ambient temperature.
2. Fill calibration containers with the buffered standards so each standard will cover the pH probe and temperature sensor.
3. Remove probe from its storage container, rinse with deionized water, and remove excess water.
4. Select measurement mode. Immerse probe into the initial standard (e.g., pH 7 buffered solution).
5. Wait until the readings stabilize. If the reading does not change within 30 seconds, select calibration mode and then select “pH”. Enter the buffered standard value into instrument.
6. Remove probe from the initial standard, rinse with deionized water, and remove excess water.
7. Immerse probe into the second standard (e.g., pH 4 buffered solution). Repeat step 5.
8. Remove probe from the second standard, rinse with deionized water, and remove excess water. If instrument only accepts two standards, the calibration is complete. Go to step ‘II. Otherwise continue. .
9. Immerse probe in third buffered standard (e.g., pH 10 buffered solution) and repeat step 5.
10. Remove probe from the third standard, rinse with deionized water, and remove excess water.
11. Select measurement mode, if not already selected. To ensure that the initial calibration standard (e.g., pH 7) has not changed, immerse the probe into the initial standard. Wait for the readings to stabilize. The reading should read the initial standard value within the manufacturer’s specifications. If not, re-calibrate the instrument. If re-calibration does not help, the calibration range may be too great. Reduce calibration range by using standards that are closer together.
12. The calibration is complete. Rinse the probe with deionized water and store the probe according to manufacturer’s instructions.
13. Record the calibration information on the calibration log sheet.
Practical Calibration Standards
Magnehelic Gauge Calibration
Turbidity Sensor Calibration Procedure
Why Calibration is Important?
DeadWeight Tester Questions
Impulse Line Pressure Testing Procedure
  • Charles Palmer says:
    April 24, 2022 at 11:55 pm
    Very comprehensive and complete pH calibration procedure, thank you.
    Just one final step : the Buffer solutions need to be discarded after the procedure, not stored for further use.
    In many Industrial plants, the 4, 7 and 10pH buffers are used again, which is not a good idea .
Control Valve Stroke Test Procedure
Difference between Calibration and Ranging
Single Position Calibration of Temperature Bath
Calibration and Troubleshooting of Oxygen Analyzer
Pressure Switch Calibration Procedure
Fieldbus Flow Transmitter Configuration
RTD Calibration
Capacitance Level Switch Calibration
Specific-Conductance-Calibration-Procedure
Specific Conductance Calibration Procedure
The Principle of Dead Weight Pressure Tester
Calibration of Weight
Calibration of Weight using the ABBA Method
Pressure Safety Valves Functional Testing
transmitter calibration procedure
Transmitter Calibration Procedure
History of Measurement
Pig Signalling Switches Functional Testing
Comparison of Accuracy and Precision
Difference between Accuracy and Precision
Industrial Oven Control Application with PLC Controller
PLC Programming Best Practices and Tips
Network Theorems Objective Questions and Answers
What is a Wet Contact in PLC
What is a Wet Contact? – Basics of PLC Wiring
Steam Pressure Taps Primary Block Valves Glands and Bonnets Leaks
Instrument Steam Pressure Taps Primary Block Valves Glands and Bonnets Leaks
desuperheater
Desuperheater Design
Single Phase Voltage Source Inverter Quiz