Reading: Magnehelic Gauge Calibration
Inst Tools > Blog > Calibration > Magnehelic Gauge Calibration
This article cover calibration of Magnehelic Gauge (Differential Pressure Gauge) using Digital Manometer.
Last updated: September 28, 2022 3:24 pm
Saurabh Vaidya
Calibration Pressure Measurement
This article cover calibration of Magnehelic Gauge (Differential Pressure Gauge) using a Digital Manometer.
Magnehelic Gauge CalibrationUp-Cycle CalibrationDown-Cycle Calibration
Magnehelic Gauge or Differential Pressure Gauge has two ports – High and Low.
As name suggest Differential Pressure Gauge i.e. its measure Pressure difference between High port and Low port.
Generally, units of Differential pressure Gauge are mmWC, Pascal, Mbar, Inch WC, PSI, mmHg
High Port is also called as Positive port and Low port is called as Negative port.
How to calculate differential pressure?
Pressure at high port is 100 Pascal and Pressure at low port is 75 Pascal.
Differential Pressure indicated by Magnehelic Gauge or Differential Pressure Gauge is 100-75= 25 Pascal,
Differential Pressure= High Port – Low Port
If pressure at Low Port is higher than Low Port, than indicator will show Negative readings.
Analog Magnehelic Gauge should be in Vertical position.
Generally, Magnehelic Gauge is installed above the door, input ports with a hose are located at the back and front of the gauge. So opening the door will equalize the ambient pressure and the Magnehelic gauge will turn back to zero (as both rooms are at equal pressure, High port – Low Port = 0)
Connect the pneumatic pump on the High Port of the Magnehelic Gauge. Leave the Low Port open to ambient, in this way, the differential pressure gauge functions as a normal pressure gauge.
 You can also connect Low Port to know Pressure and find the difference between High and low port.
The calibration Procedure from this step is the same as the Pressure Gauge.  I.e. Preloading, UP Cycle, Down Cycle.
Equally distribute calibration points into 5 points (0%, 25%, 50 %, 75% and 100%)
Pre Loading – Increase Pressure up to full scale & decrease it to zero of UUC, to make pointer movement free.
We should wait at least 30 seconds to note the readings of Magnehelic Gauge. The waiting time for the upper limit of the measuring range is 2 minutes.
Up-Cycle is measured from Minimum pressure range to Maximum Pressure Range
Down-Cycle is measured from Maximum pressure range to Minimum Pressure Range
  • Increase Pressure by operating the pump to set the calibration point on the UUC gauge.
  • When pointer of UUC Gauge reach first calibration point note down both UUC & standard reading. Wait for 30 seconds to note down reading. Wait for 2 Minutes for upper limit of measuring range.
  • Slowly increase the pressure to set next set point .Write down reading Standard reading & UUC reading.
  • Repeat this procedure for remaining calibration points.
  • This completes Up cycle of calibration of Magnehelic Gauge
  • For Down Cycle, Last calibration point of Up Cycle is first point of Down Cycle.
  • Slowly decrease the pressure to set the next set point. Write down reading Standard reading & UUC reading. Wait for 30 seconds to note down reading. Wait for 2 Minutes for the upper limit of the measuring range.
  • This completes the Down Cycle of calibration of Magnehelic Gauge
  • History of Measurement
  • Pressure Gauge Calibration
  • Selection of Master Instrument
  • Importance of Unit Measurement
  • Reliability Centered Maintenance
Prover Tank Calibration and Mass Flow Meter Verification Procedure
Instrument Errors – Zero, Span, Linearity
Pressure Gauge Syphons Principle
Functions of Pressure Detectors
Pressure Gauges : Remote and Chemical Seals
Pressure Gauge Selection
RTD Calibration
Instrument Calibration Lab Exercise
Instrumentation Engineer’s Calibration Mistakes
Calibration and Troubleshooting of Oxygen Analyzer
H & B Gas Analyzer Principle and Calibration Procedure
Closed tank DP Level Transmitter with wet leg Zero elevation Calibration
DP Diaphragm Capillary Seal Level instrument Calibration Procedure
Vacuum Gauges Working Principle
Capacitance Pressure Sensor Works
How Capacitive Pressure Sensor Works ?
Calibration & Preventative Maintenance Procedures
Calibration and Preventative Maintenance Procedures
Hammer Effect in Gauges
What is Hammer Effect in Gauges ?
How to Test a Pressure Switch ?
Gas Detectors Bump Testing
What is Bump Testing ?
Portable Gas Detectors Calibration Procedure
Pressure Gauge Water traps and pigtail siphons
Pressure Gauge Water Traps and Pigtail Siphons
DeadWeight Tester
DeadWeight Tester Questions
Dome Valve Principle
Dome Valve Working Principle
Relay Parts
NO vs NC Contact
Classification of Flow Measurement Techniques
RTD-standards
How to calculate the tolerance of a RTD sensor
SCADA Standards
SCADA for Substation Automation
Electrical Machines MCQ Series 18