Suppose an electronic pressure transmitter has an input range of 0 to 100 PSI and an output range of 4 to 20 mA. When subjected to a 5-step up-and-down “As-Found” calibration test, it responds as such:
Sketch this instrument’s ideal transfer function on the graph, along with its actual transfer function graph based on the measured values recorded as shown in above table. Then, determine what kind of calibration error it has (zero shift, span shift, hysteresis, and/or linearity):
Finally, identify how this calibration error might be corrected. What steps or procedures would you follow to rectify this problem?
Solution:
This instrument has a zero shift error, but not a span shift or linearity error.
Ideal transfer function
Read: Instrument Calibration Errors
Actual Transfer Function
Zero Error:
Span Error:
A span error would look something like this (wrong slope):
Linearity Error:
A linearity error would look something like this (not a straight line):
A zero error is usually correctable by simply adjusting the “zero” screw on an analog instrument, without making any other adjustments.
Span errors, by contrast, usually require multiple adjustments of the “zero” and “span” screws while alternately applying 0% and 100% input range values to check for correspondence at both ends of the linear function.
For a SMART Transmitter, you can do zero trim and span trim using a HART communicator, by applying the respective value of process variable to the transmitter input and apply zero & span trims accordingly.
Answer the Questions via comments:
- How might the other three calibration errors appear when graphed?
- What purpose is served by doing an up-and-down test? Why not just check the instrument’s response in one direction only?
- Which constant in the y = mx + b linear equation represents zero, and which represents span?
Credits: Tony R. Kuphaldt
- Instrument Turndown Ratio
- Tunning of PID Controller
- 4-20 mA Formula
- Transmitter LRV & URV
- Control Loop Test