Reading: Shunt Calibration of a Strain Gauge Load Cell
Inst Tools > Blog > Calibration > Shunt Calibration of a Strain Gauge Load Cell
Last updated: August 22, 2019 7:44 am
Calibration
Shunt calibration is the known electrical unbalancing of a strain gauge bridge by means of a fixed resistor that is placed, or “shunted,” across one leg of the bridge.
The Wheatstone Bridge used in strain gauge load cells are typically field-calibrated using the shunt calibration technique.
Shunt calibration is a method of periodically checking the gain or span of a signal conditioner, which is used in conjunction with a strain gauge based transducer, without exposing the transducer to known, traceable, physical input values.
If required, adjustments can then be made to the signal conditioner to insure accurate measurement results
A strain gauge bridge is “in balance” when the host mechanical structure is unloaded and unstressed.
As the host structure (diaphragm, bending beam, shear beam, column, etc.) is loaded or stressed, the Wheatstone Bridge becomes unbalanced, resulting in an output signal that is proportional to the applied load.
Calibration of a Strain Gage Load Cell
Fig : A Wheatstone Bridge circuit showing the location for connecting the appropriate shunt resistor for the purpose of simulating either a tension or compression input.
Shunt calibration stimulates the mechanical input to a transducer by unbalancing the bridge with a fixed resistor placed across, or in parallel with, one leg of the bridge.
For tension shunt calibration, the shunt resistor (RST) is shunted across the +excitation (+P) and +signal (+S) leg of the bridge.
For a compression shunt calibration, the shunt resistor (RSC) is shunted across the – excitation (-P) and +signal (+S) leg of the bridge.
Calibration Procedure :
Here are step-by-step instructions for shunt calibration of a strain gauge load cell.
1. Connect the transducer to an appropriate strain gauge signal conditioner and allow adequate time for the system to stabilize.
2. Apply a full-scale N.I.S.T. traceable, mechanical input (or load) to the transducer.
3. Adjust the signal conditioner’s gain or span controls, as required, to obtain a full-scale electrical output signal, and/or numeric display that represents the applied, mechanical input quantity.
4. Remove the mechanical input (or load).
5. Place a shunt calibration resistor across an appropriate leg of the Wheatstone Bridge (as discussed above).
6. Record the value of the signal conditioner’s output signal and/or numeric display. This value is the shunt calibration value, or equivalent load.
7. It is important to note that the shunt calibration value is specific for the particular shunt resistor used.This value, and the particular resistor, are now matched to the transducer and form the basis of the transferable shunt calibration.
Summary
Shunt calibration is accepted throughout the industry as means of periodic calibration of a signal conditioner and transducer between calibrations of known, applied, traceable, mechanical, input values.
Consequently, most all strain gauge transducer manufacturers collect and supply shunt calibration data, along with a shunt calibration resistor, as a standard feature.
Source : PCB Load & Torque Division
Fieldbus Flow Transmitter Configuration
Conductivity Transmitter Calibration Procedure
Instrument Maintenance and Calibration System
Common Mistakes to Avoid When Using Dead Weight Tester
Differential Pressure Switch Calibration Procedure
Different Types of Temperature Calibrators
Selection of Weighing Balance and Weight for Calibration
Oxidation-Reduction Potential (ORP) Sensor Calibration Procedure
Stability and Uniformity of Temperature Bath
Prover Tank Calibration and Mass Flow Meter Verification Procedure
How to Calibrate the Eddy Current Conductivity Meter?
Portable Gas Detectors Calibration Procedure
Step by Step Guide for Gas Flow Meter Calibration
Selection of Industrial Temperature Calibrator
Changing Process Medium Density has no influence on Level Indication (Full Absorption)
List of Documents Required for Testing and Calibration Laboratory
List of Documents Required for Testing and Calibration Laboratories
Fieldbus Transmitter Calibration
differential-pressure-transmitter-calibration-procedure
Differential Pressure Transmitter Calibration Procedure
Instrumentation Engineer’s Calibration Mistakes
DP Diaphragm Capillary Seal Level instrument
DP Diaphragm Capillary Seal Level instrument Calibration Procedure
level-transmitter-calibration-procedure
Level Transmitter Calibration Procedure
What is a Magnetic Contactor
What is a Magnetic Contactor? – Principle, Types, Advantages
controller integral wind-up
Limit Controls : High Limit & Low Limit Functions
AC Voltage Controllers Objective Questions
Safety Valve
Compare Relief Valve and Safety Valve
6 Pieces of Essential Equipment For PCB Assembly
Design a PLC program for Operational Sequence of Packaging Process
Piston Check Valve
What is Piston Check Valve ?