The following empirically derived thermocouple laws, are useful to understand, diagnose and utilise thermocouples.
a) Law of homogeneous circuits
If two thermocouple junctions are at T1 and T2, then the thermal emf generated is independent and unaffected by any temperature distribution along the wires.
In above Figure, a thermocouple is shown with junction temperatures at T1 and T2. Along the thermocouple wires, the temperature is T3 and T4. The thermocouple emf is, however, still a function of only the temperature gradient T2 – T1.
b) Law of intermediate metals
The law of intermediate metals states that a third metal may be inserted into a thermocouple system without affecting the emf generated, if, and only if, the junctions with the third metal are kept at the same temperature.
Law of intermediate metals Principle
When thermocouples are used, it is usually necessary to introduce additional metals into the circuit This happens when an instrument is used to measure the emf, and when the junction is soldered or welded. It would seem that the introduction of other metals would modify the emf developed by the thermocouple and destroy its calibration. However, the law of intermediate metals states that the introduction of a third metal into the circuit will have no effect upon the emf generated so long as the junctions of the third metal are at the same temperature, as shown in Above Figure
c) Law of intermediate temperatures
The law of intermediate temperatures states that the sum of the emf developed by a thermocouple with its junctions at temperatures T1 and T2, and with its junctions at temperatures T2 and T3, will be the same as the emf developed if the thermocouple junctions are at temperatures T1 and T3.
Law of intermediate temperatures Principle
This law, illustrated in above Figure, is useful in practice because it helps in giving a suitable correction in case a reference junction temperature other than 0 °C is employed. For example, if a thermocouple is calibrated for a reference junction temperature of 0 °C and used with a junction temperature of 20 °C, then the correction required for the observation would be the emf produced by the thermocouple between 0 °C and 20 °C.
Also Read: RTD or Thermocouple ?
Thermocouple Calibration
Resistance Temperature Detectors Troubleshooting Tips
Relationship between Temperature Scales
Thermocouples Green Rot Effect
Optical Pyrometer Working Principle
Resistance Temperature Detectors (RTD) Construction
Thermowell Wake Frequency Calculation
Calculate Temperature Transmitter Output
Multiple Position Calibration of Temperature Chamber
Bi-metal Temperature Sensors
Temperature Sensor Failure
How a 4 Wire RTD Works ?
What is a Thermocouple ?
Liquid In Glass Thermometers
Liquid In Glass Thermometer Principle
Thermocouple Wiring Diagram
Basics of Thermocouples and RTD
Difference Between RTD, Thermocuples and Thermistors
Difference Between RTD, Thermocouple and Thermistor
Functions of Temperature Detectors
Measuring temperature with RTD
How RTD Measuring the Temperature
Calibration of Temperature Sensor with Indicator
Temperature Sensors Fault
Skin Type Thermocouple
What is Open Telemetry?
What is Open Telemetry? – Principles and Benefits
Fundamental Motion Control Commands
Valve Standards
List of Valve Standards
Current in a Capacitor
Capacitance and Capacitive Reactance
Transformer Efficiency
Rotary Encoder
Rotary Encoder Interview Questions and Answers
Valve Basics, Sizing, and Actuators Books
What is Trunking
Difference Between Trunking and Conduit