Inst ToolsInst ToolsInst Tools
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Search
  • Books
  • Software
  • Projects
  • Process
  • Tools
  • Basics
  • Formula
  • Power Plant
  • Root Cause Analysis
  • Electrical Basics
  • Animation
  • Standards
  • 4-20 mA Course
  • Siemens PLC Course
Reading: Dissolved Oxygen Analyzer Working Principle
Share
Font ResizerAa
Inst ToolsInst Tools
Font ResizerAa
  • Courses
  • Design
  • PLC
  • Interview
  • Control System
Search
  • Courses
  • Automation
    • PLC
    • Control System
    • Safety System
    • Communication
    • Fire & Gas System
  • Instrumentation
    • Design
    • Pressure
    • Temperature
    • Flow
    • Level
    • Vibration
    • Analyzer
    • Control Valve
    • Switch
    • Calibration
    • Erection & Commissioning
  • Interview
    • Instrumentation
    • Electrical
    • Electronics
    • Practical
  • Q&A
    • Instrumentation
    • Control System
    • Electrical
    • Electronics
    • Analog Electronics
    • Digital Electronics
    • Power Electronics
    • Microprocessor
  • Request
Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Inst Tools > Blog > Analyzers > Dissolved Oxygen Analyzer Working Principle

Dissolved Oxygen Analyzer Working Principle

Last updated: November 11, 2017 12:41 pm
Editorial Staff
Analyzers
No Comments
Share
5 Min Read
SHARE

What is dissolved oxygen ?

Dissolved oxygen refers to oxygen dissolved in water. Its concentration is expressed as the amount of oxygen per unit volume and the unit is mg/L. Biologically, oxygen is an essential element for respiration of underwater life and also acts as a chemical oxidizer. The solubility of oxygen in water is affected by water temperature, salinity, barometric pressure, etc. and decreases as water temperature rises.

Measurement of dissolved oxygen by the membrane electrode method

The membrane electrode method measures a diffusion current or reduction current generated by the concentration of dissolved oxygen or partial pressure of oxygen to obtain the concentration of dissolved oxygen. This method is not affected by the pH value of water being measured, oxidation and reduction substances, color, turbidity, etc. and the measurement method offers good reproducibility.

When a sensor is inserted into water, an air layer forms on the membrane (Teflon membrane). The oxygen partial pressure (concentration) in the air layer is in equilibrium with the concentration of dissolved oxygen in the water. The membrane electrode method measures the oxygen concentration in the gas phase to indirectly obtain the concentration of dissolved oxygen in water.

There are two types of membrane electrode method:

Galvanic cell method, and

Polarographic method.

These methods differ only in the presence or absence of an external applied voltage and have the same performance, features, and usage method.

(1) Galvanic cell method

The membrane has high permeability to oxygen and is constructed so that the electrodes and electrolyte are isolated from the water being measured. The counter electrode is a base metal and the working electrode is a noble metal and potassium hydroxide is used as the electrolyte. Oxygen passes through the membrane and is reduced on the working electrode, and so the method measures the reduction current flowing between both electrodes, which is proportional to the concentration of dissolved oxygen.

dissolved oxygen analyzer principle

(2) Polarographic method

The sensor construction is almost the same as that of the galvanic cell method. The counter electrode is silver-silver chloride and the working electrode is gold or platinum. When a voltage of 0.5–0.8 V is applied between both electrodes, oxygen that has permeated through the membrane initiates a reduction reaction on the working electrode, causing a polarographic limiting current to flow which is proportional to the oxygen concentration. This method measures the concentration of dissolved oxygen based on this current value.

DO Analyzer Polarographic method

Calibration of the dissolved oxygen analyzer

Calibration of the dissolved oxygen analyzer is performed in the following situations:

– When a new dissolved oxygen sensor is installed
– When the membrane assembly is replaced and/or the electrolyte solution is replaced
– When the sensor has been disassembled and reassembled for maintenance
– When a measuring error after cleaning exceeds the acceptable deviation from the reference method

There are the following methods for calibrating the dissolved oxygen analyzer:

Air calibration:
Performs span calibration in ambient air. This is the most common and easiest way to calibrate the analyzing system, and the method is generally satisfactory. The calibration table lists experimental values at a relative humidity of 70%. It is important to wash any contamination from the sensor membrane, and then wipe off remaining water droplets from the membrane gently with soft paper, etc.
Span calibration in air-saturated water:
This method complies with ISO 5814; perform this calibration if you wish to calibrate the analyzer more accurately.
Zero calibration in sodium sulfite solution:
This calibration takes much time, and is not generally required.
Real solution calibration: Measurement is made using an instrument based on the water being measured and calibration is performed in accordance with the value obtained.

Also Read: Conductivity Analyzer Interview Questions & Answers

Don't Miss Our Updates
Be the first to get exclusive content straight to your email.
We promise not to spam you. You can unsubscribe at any time.
Invalid email address
You've successfully subscribed !

Continue Reading

Infrared Non Dispersive CO2 Analyzer Working Principle
TOC Analyzer Working Principle
Two Electrode Conductivity Probes Principle
Fluorescence
Troubleshooting pH Analyzer Common Problems
Conductivity Measurement
Share This Article
Facebook Whatsapp Whatsapp LinkedIn Copy Link
Share
Leave a Comment

Leave a Reply Cancel reply

Your email address will not be published. Required fields are marked *

Stay Connected

128.3kFollowersLike
69.1kFollowersFollow
210kSubscribersSubscribe
38kFollowersFollow

Categories

Explore More

Top 100 Analytical Instrumentation Engineering Projects
Gas Filter Correlation (GFC) Spectroscopy
What is Gas Chromatography?
Online Gas Chromatograph Principle
Dissolved oxygen probe calibration procedure
Falling Ball Viscometer Principle
What is Gas Metering System?
Moisture Calculator

Keep Learning

Suspended Particulate Matter

Basics of Suspended Particulate Matter (SPM) Analyzers

Various Types of Sensors Used in Water Treatment Plant

Various Types of Sensors used in Water Treatment Plant

Total Organic Carbon Analyzers Questions and Answers

Total Organic Carbon Analyzers Questions and Answers

mass spectrometer sample system

Mass Spectrometer Working Principle

Dissolved Oxygen Analyzers Interview Questions

Dissolved Oxygen Analyzer Interview Questions & Answers

Turbidity Analyzer Working Principle

Turbidity Analyzer Working Principle

online silica analyzer

Silica Analyzer Problems and Troubleshooting Tips

Pulsed Fluorescence SO2-H2S-CS Analyzer Working Principle

Pulsed Fluorescence SO2, H2S, CS Analyzer Working Principle

Learn More

Determine Calibration Range of Pneumatic Pressure Transmitter

Determine Calibration Range of Pneumatic Pressure Transmitter

water filter level control system

Practical Process Control System Questions & Answers – 6

Difference Between PLC and CNC Machine

Difference Between PLC and CNC Machine

Introduction to Mechatronics

Introduction to Mechatronics

Why we use Cable Shielding with Ferrites ?

Why we use Cable Shielding with Ferrites ?

Annunciators - 2

What is a Annunciator ?

Electropneumatic Valve Positioner Working Principle

Electro Pneumatic Valve Positioner Schematic & Principle

Control Valve Whisper Trim Cage

Problem with Choked Control Valve with Whisper Trim Cage

Menu

  • About
  • Privacy Policy
  • Copyright

Quick Links

  • Learn PLC
  • Helping Hand
  • Part Time Job

YouTube Subscribe

Follow US
All rights reserved. Reproduction in whole or in part without written permission is prohibited.
Welcome Back!

Sign in to your account

Username or Email Address
Password

Lost your password?